ICD Dual Chamber Generator Change
Indications
ICD (implantable cardioverter-defibrillator) battery depletion [Z45.02 (ICD-10-CM)]
Pre Procedure Diagnosis
ICD (implantable cardioverter-defibrillator) battery depletion [Z45.02]

Conclusion
Procedure:
Generator Change out: ICD
Indication: Device at ERI
Sedation: Please see anesthesia record
Antibiotics: Ancef and vancomycin IV

The patient was prepped in the usual way for CRM generator change out

Time-out performed

The existing incision was cut using the blade and electrocautery.

Using electrocautery and blunt dissection in the subcutaneous fascia the device pocket was entered. The capsule was entered.

The generator was removed from the pocket. The existing cables were detached from the header and attached to the new pulse generator header.


The pocket was flushed copiously with antibiotic solution.


The pulse generator was placed inside the pocket with the cables coiled underneath it. A tyrx pouch was used.

The pocket was then closed in the usual fashion using 2-0 Vicryl, 3-0 and 4-0 Vicryl.

Steri-Strips were applied over the incision.


A light dressing was placed over the incision. The patient was transferred to a monitored bed in stable condition.

Company Model Serial number
Pulse generator Medtronic
Atrial lead Medtronic
Right ventricular lead Medtronic
LV Lead Medtronic
Explanted Pulse Generatpr Medtronic


Right ventricle Left Ventricle Right atrium
Threshold 0.75 volts at 0.4 milliseconds (1-2)
1 at 0.4 -
Sensing 11.6mV 2.5 mV
Impedance 418 Ohms 589 456 Ohms


VF 200, ATP during charge, max shocks
Monitor 182

VVIR 70-120

Conclusion:

Successful BIVICD generator change out
D/C today
Oral Abx x 7 days
F/u in device clinic as scheduled
 
Hello for the change of generator (battery only) is 33264 you know it is a multiple lead system due to the testing at bottom states RV, RT atrium, and LV.

There are 4 chambers of the heart RV, Rt atrium on the Rt side and LV and Lt atrium on the left side.
Since the previous leads are in the RV, Rt atrium and LV this is a multiple lead system for ICD.
Where as a dual lead is two chambers of the heart and of course single lead is one chamber (all of these can be either in heart or subcutaneous electrodes).
Info gathered from CPT book 2023 pg 236-237
 
Hello for the change of generator (battery only) is 33264 you know it is a multiple lead system due to the testing at bottom states RV, RT atrium, and LV.

There are 4 chambers of the heart RV, Rt atrium on the Rt side and LV and Lt atrium on the left side.
Since the previous leads are in the RV, Rt atrium and LV this is a multiple lead system for ICD.
Where as a dual lead is two chambers of the heart and of course single lead is one chamber (all of these can be either in heart or subcutaneous electrodes).
Info gathered from CPT book 2023 pg 236-237
Thank you for your input... it's the same as what the auditor told me. I had used 33263.
 